Appendix A.Specifications

A.1Strain Gauge

The table below lists the specifications for all strain gauge models.

Model

4200
4200L

4200 6
4200
 7

4202

4204

4210

4212

4214

4200HT/HT-T

Range (Nominal)

3000 με

5,000 με
10,000 με

3000 με

Resolution

1.0 με1

2 − 5 με1

0.4 με1

1.0 με1

0.4 με1

0.4 με1

0.4 με1

1.0 με1

Calibration Accuracy

±0.5% Full Scale Reading (FSR)

Batch Factor Accuracy

±0.5% FSR

2.5 − 4% FSR

±0.5% FSR

System Accuracy

±2.0% FSR2

Stability

0.1% Full Scale (FS)/yr

Linearity

0.5% FSR

2.5 − 4% FSR

0.5% FSR

Thermal Coefficient3

12.2 ppm/°C3

12.2 ppm/°C3
17.3 ppm/°C3

Frequency Range (Hz)

450 − 1200

450 − 1200

1400 − 3500

800 − 1600

1400 − 3500

1400 − 3500

1400 − 3500

450 − 1200

Dimensions (gauge) (LxD)

6.125 x 0.750"
155 x 19 mm

6.00 x 0.750"

153 x 19 mm

2.250 x 0.625"

57 x 16 mm

4.125 x 0.750"

105 x 19 mm

10.250 x 2"

260 x 50 mm

12.250 x 2"

311 x 50 mm

14.250 x 2"

362 x 50 mm

6.125 x 0.750"

155 x 19 mm

Dimensions (coil)

0.875 x 0.875"

22 x 22 mm

0.875 x 0.875"

22 x 22 mm

N/A

0.875 x 0.875"

22 x 22 mm

Coil Resistance

180Ω

180Ω

50Ω

180Ω

180Ω

180Ω

180Ω

120Ω / 50Ω

Temperature Range

20 to +80 °C

−20 to +200 °C
−20 to +220 °C

 

table 4: Strain Gauge Specifications

Notes:

1 Possible resolution depends on the readout; the figures in the above table pertain to the GK-404.

2 System Accuracy takes into account hysteresis, nonlinearity, misalignment, batch factor variations, and other aspects of the actual measurement program. System Accuracy to ±1.0% FS may be achieved through individual calibration of each strain gauge.

A.2Thermistor

For more information, refer to Appendix C.

Range: 80 to +150 °C

Accuracy: ±0.5 °C