Model 6150F inclinometers use the industry standard Modbus Remote Terminal Unit (RTU) protocol to communicate with the chosen readout method. As the name suggests, Modbus was designed to work on what is known as a bus network, meaning that every device receives every message that passes across the network. Model 6150F inclinometers use the RS-485 electrical interface because of its prevalence, simplicity, and success as a robust, industrial physical layer.
More information about Modbus can be found at the following website:
http://www.modbus.org/specs.php
The Modbus RTU protocol uses packets (messages made up of multiple sections) to communicate and transfer data between devices on the network. The general format of these packets is as follows:
1.Modbus Address (one byte) – The address of the specific device on the bus. (Labeled on the sensors as #1, #2, #3, etc.)
2.Function Code (one byte) – The action to be carried out by the slave device.
3.Data (multi-byte) – The payload of the function code being sent.
4.Cyclic Redundancy Check or CRC (two bytes): A 16-bit data integrity check calculated over the other byes in the packet.
The most recent sensor readings are stored in memory registers, read using a Modbus command. Angle and temperature readings are available in processed or precursor formats. Register addresses and formats are described in Table 2.
Table 3 shows device control addresses. Any nonzero value written to the trigger address initiates a measurement cycle, updating the angle and temperature measurement registers. Any anomalies detected during the most recent measurement cycle produce a non-zero error code. Refer to Appendix C for an explanation of these codes.
The flash password prevents unintended writes to the nonvolatile memory in Table 4 and the preprogrammed device information in Table 5. Contact geokon for instructions.
Byte |
Word |
Parameter |
Units |
Type |
Access |
|
---|---|---|---|---|---|---|
0x100 |
0 |
LSW |
A-Axis |
degrees |
float |
RO |
1 |
||||||
0x101 |
2 |
MSW |
||||
3 |
||||||
0x102 |
4 |
LSW |
B-Axis |
degrees |
float |
|
5 |
||||||
0x103 |
6 |
MSW |
||||
7 |
||||||
0x106 |
12 |
LSW |
Temperature |
°C |
float |
|
13 |
||||||
0x107 |
14 |
MSW |
||||
15 |
||||||
0x108 |
16 |
LSW |
Uncorrected A-Axis |
degrees |
float |
|
17 |
||||||
0x109 |
18 |
MSW |
||||
19 |
||||||
0x10A |
20 |
LSW |
Uncorrected B-Axis |
degrees |
float |
|
21 |
||||||
0x10B |
22 |
MSW |
||||
23 |
||||||
0x10E |
28 |
LSW |
Thermistor ADC |
N/A |
uint16 |
|
29 |
||||||
0x117 |
46 |
|
Error Code |
N/A |
uint16 |
|
47 |
table 2: Register Addresses and Formats
Byte |
Word |
Parameter |
Units |
Type |
Access |
|
---|---|---|---|---|---|---|
0x118 |
48 |
|
Trigger |
N/A |
uint16 |
RW |
49 |
||||||
0x119 |
50 |
LSW |
Password |
N/A |
uint32 |
|
51 |
||||||
0x11A |
52 |
MSW |
||||
53 |
||||||
0x11B |
54 |
|
Measure Cycle |
N/A |
uint16 |
|
55 |
table 3: Device Control Addresses
Byte |
Word |
Parameter |
Units |
Type |
Access |
|
0x200 |
0 |
|
Drop Address |
N/A |
uint16 |
RO |
1 |
||||||
0x201 |
2 |
|
Sensor Type |
N/A |
string |
|
3 |
||||||
0x202 |
4 |
|||||
5 |
||||||
0x203 |
6 |
|||||
7 |
||||||
0x204 |
8 |
|||||
9 |
||||||
0x205 |
10 |
|||||
11 |
||||||
0x206 |
12 |
|||||
13 |
||||||
0x207 |
14 |
|||||
15 |
||||||
0x208 |
16 |
|||||
17 |
||||||
0x209 |
18 |
LSW |
Serial Number |
N/A |
uint32 |
|
19 |
||||||
0x20A |
20 |
MSW |
||||
21 |
||||||
0x20B |
22 |
|
Software Version |
N/A |
uint16 |
|
23 |
||||||
0x20C |
24 |
|
Hardware Version |
N/A |
uint16 |
|
25 |
Byte |
Word |
Parameter |
Units |
Type |
Access |
|
0x20D |
26 |
LSW |
A Offset |
degrees |
float |
RO |
27 |
||||||
0x20E |
28 |
MSW |
||||
29 |
||||||
0x20F |
30 |
LSW |
B Offset |
degrees |
float |
|
31 |
||||||
0x210 |
32 |
MSW |
||||
33 |
||||||
0x213 |
38 |
LSW |
A Gauge Factor |
degrees |
float |
|
39 |
||||||
0x214 |
40 |
MSW |
||||
41 |
||||||
0x215 |
42 |
LSW |
B Gauge Factor |
degrees |
float |
|
43 |
||||||
0x216 |
44 |
MSW |
||||
45 |
|
table 5: Preprogrammed Device Information